Topic: High Conductivity
|
Greg unregistered
|
posted 02-02-2003 01:11 PM
m micone, I have not said that any particular thing is what's causing my tx problems. That is what I'm trying to find out. 1) I do not cramp at all as I set my goal correctly each tx. What I said is, I feel crampy when I have been put on machines that register a conductivity of below 14.3. 2) Before the high cond. alarm problem on my regular machine, my KT/Vs were totally steady for years. Now they are up and down. 3) I am also on Renagel and my CO2 began declining several months ago. I was put on sodium bicarbonate and CO2 is in limits again. 4)If the highest bicarb you can get on the machne is 37, does that say you are on a one size fits all bicarb mix that might not be approripate for you? 5)The high cond alarm on my machine does not go off only when they change out the jug - it goes off repeatedly most txs (as many as 6-8 times) and some txs does not go off at all. 5) I would like to use sodium modeling, but my staff is not well trained in it and I don't want to be a guniea pig. However, as I said, I don't need it to prevent cramping, because I never cramp. If I used it, it would be to provide a smoother tx.
IP: 209.240.198.61 |
Chuck W Member
|
posted 02-02-2003 03:09 PM
Greg,The "one size fits all" bicarb level that you refer to can not be gotten around. The levels that are obtainable are determined by the proportioning ratio of the machine. As an example, you can not use a 44:1 bicarb with a 34:1 acid, to do so would drive the conductivity through the roof.
IP: 205.188.193.153 |
Enough unregistered
|
posted 02-02-2003 08:14 PM
Enough already!!!
IP: 67.26.53.216 |
mgunsolus Member
|
posted 02-03-2003 01:57 PM
Greg,You shouldn't need to talk with the machine tech. If they make bicarb and acid on site any tech should be able to tell you what the specific gravity is. I wouldn't think that the machine tech is the only one to make bicarb and acid. Is the machine tech at your clinic also the cheif tech? Can you not DEMAND the MD and unit manager make an appointment to talk with the tech? I know I sure would! I have patients ask to speak with me all the time and if I can't give the anwser they want I will get the anwser for them. Is this guy or gal just that mean tthat they will not talk with you? Affraid you will stump him? Anyways enough rambling. Good Luck and I will Check back in a week or so.
IP: 208.46.13.226 |
Smiler Member
|
posted 02-03-2003 03:52 PM
quote: Originally posted by m micone: I'm a patient, on a fresenius 2008h. I commisserate, but you may be blaming symptoms on the wrong culprit. Cramping is a fluid removal question. Appetite can relate to many things, but it's most often an adequcy of dialysis question. I've been having a problem with low co2 due to using renagel. We tried riasing the bicarb proportioning on the machine settings to compensate. The highest we can et is 37, due to the bicarb mix used on all the machines. If you set it higher, you get no result. The only time our machines alarm for high conductivity is when the techs change to a full bicarb jug in mid tx and the machine needs a moment to adjust to the fresh batch. For cramping, if you are taking off more than two litres, ask about sodium modeling scripts. Hope this helps.
Using Naturalyte 9000 series concentrates, the machine should deliver up to 40 mEq/. This should be verified by sending dialysate samples periodically for electrolyte and bicarbonate composition. We have verified bicarb reults up to 40 mEq/l and as low as the usual standard of 35 mEq/l.
IP: 4.17.129.194 |
Greg unregistered
|
posted 02-04-2003 03:04 PM
jtech amd mgunsoius,I asked several staff about the bicarb proportioning and specific gravity. They didn't seem to know these two terms at all and did not seem to know who to refer me to. I finally located the person who mixes the bicarb. My unit has a maintenance person doing the job. She gave me the product insert for Naturalyte 4000. It says 1 part acid to 44 parts bicarbonate base and water is used. Nominal dilution is 1:1.72:42.28. Sodium =37mEq/L. Bicarbonate = 37mEq/L. Chemical Composition: 650gm. Sodium B, Bicarbonate. U.S.P. Dissolved 81.25. Next I will ask this person if she can tell me the specific gravity. Where is the info found about the proportioning of the machine? What do you mean by proportioning of the machine? It says on the bicarb insert: "Check conductivity and pH of dialyzing fluid before starting tx and each time solution is added (AAMI RD5)". Conductivity and pH is checked in my unit with strips prior to the tx, but it is not checked when the jugs run out and new bicarb is added.
IP: 209.240.198.61 |
mgunsolus Member
|
posted 02-04-2003 04:47 PM
Hey Greg, I don't know of a strip that can check the conductivity of the machine. Also how is the bicarb mixed 50 gallons at a time etc... Is the 4000 the only kind they use?
IP: 208.46.13.226 |
Tailor unregistered
|
posted 02-04-2003 09:37 PM
Greg, the proportioning people refer to is the amounts of acid, bicarb and water used to make the dialysate. You are using what is called "Cobe" or "Standard Cobe" proportioning, and it consists of: 1 part Acid 1.72 parts bicarb 42.28 parts water Just like mixing a martini
IP: 167.83.10.20 |
In a position to know unregistered
|
posted 02-05-2003 07:16 AM
Greg, or Jane, or patient, or whoever you are: (remember your IP address is displayed) You don't appear to be interested in resolution, but perpetuating debate. The conductivity issues you describe, and the sypmtoms you describe are challenging credibility. In 22 years I'm at a loss to recall anything like your issue and I go back to the days of acetate dialysis. Before you protest, I won't exchange ideas based on scientific method with someone lacking the training and experience. Despite information given to you by posts on this forum, you keep repeating the same behavior over and over, expecting a diferent result.
IP: 209.244.215.250 |
Greg unregistered
|
posted 02-05-2003 02:05 PM
mgunoius, I meant prior to each tx the pH is checked with a strip. (in other units I have been in pH was checked with a meter). pH and conductivity are then notated on the run sheet. pH is always notated as 7.0, but cond. varies. Previously, a tech told me bicarb is mixed in a large tub. The instructions for Naturalyte 4000 only mention mixing for 2.1 gallon jugs. Tailor, Is the proprtioning you described set by the machine?
IP: 209.240.198.61 |
m micone Member
|
posted 02-05-2003 03:49 PM
SMILER I'm at my unit right now. Are your specs for naturalite a definite result. We use the 6000 on 2008H. Where can I find more tech info. m micone
IP: 4.23.90.6 |
mgunsolus Member
|
posted 02-05-2003 04:12 PM
Greg,OK I'm sorry I miss understood about the strips. Now for the bicarb: Is it mixed in the jugs or is it mixed in a bulk tank?Either way here could be one of the problems. If they are mixing it in the jugs they could be adding to little water for it to mix properly. If they are mixing it in a bulk tank ie 50 gallons at a time this could be the problem as well to little water. In my opinion it is really unsafe to mix bulk bicarb using these little boxes. I know we used to use the for our Acute unit. Not to mention that it takes alot of time to open all of those little cartons. I wouldn't want to be the one doing it. It also could be that the person making it is using enough water but is adding to much bicarb. How long has this problem been going on? Is it just in the winter months? If so it could be that the water used to make the bicarb is too cold and may need to be mixed longer. I know at my clinic this is the case we have to mix a little longer in winter than we do in the spring and summer.
IP: 208.46.13.226 |
Greg unregistered
|
posted 02-05-2003 04:54 PM
m micone, What pH are you getting?mgunsoius, I will ask next time I go to dialysis how the bicarb is mixed. It may be that the little boxes are not used when they make the bicarb and it was just that I was given an insert from a box to answer my question about the proportioning. Also, remember, no one is having a problem with high conductivity alarms but me. The problem started about 6 months ago.
IP: 209.240.198.61 |
GTSCSS Member
|
posted 02-05-2003 05:18 PM
If it's only YOUR machine that alarms then maybe the staff should have the machine pulled and functionally tested by a technician? Also, they shouldn't be stopping that frequently to 'mute' conductivity alarms. They should report it as a machine problem and change it out. Also, request that a dialysate sample (of the dialysate you use on your treatment) be sent off to a lab to verify it's composition. quote: Originally posted by Greg: jtech, I asked the techs and RNs what the proportioning of the machine is and none of them could answer the question. I also asked about the proportioning of the bicarb and no luck there either. I won't be able to answer these questions until I can talk to the machine tech. I would prefer using factory mixed solutions as I think they're safer, but I doubt that I could request it as it wouldn't be cost effective.mgunsoius, Likewise, unable to find out about the hydrometer measurement of specific gravity until I can talk to the machine tech. Anxiety, I follow the diet very carefully and have excellant labs each month. I have never skipped or shortened a tx in years of dialysis. I understand about the clocks and I am the one who insists the techs not end the tx until the RTD clock has elapsed. No other patient's machine has high cond. alarms. The problem is with my machine, alone. My RNs and techs can tell you that, because it makes their job a whole lot harder as they always have to stop what they are doing to silence the alarms on my machine.
[This message has been edited by GTSCSS (edited 02-05-2003).]
IP: 65.215.145.67 |
Smiler Member
|
posted 02-06-2003 04:43 PM
quote: Originally posted by m micone: SMILER I'm at my unit right now. Are your specs for naturalite a definite result. We use the 6000 on 2008H. Where can I find more tech info. m micone
We sample three machines randomly every two weeks, at different bicarb setpoints. The samples are verified by lab analysis. For information on Naturalyte 9000 series concentrates, contact whomever it is you purchase the 6000 series from . The effective yield is 35 mEq/L which can be adjusted as high as 40 mEq/L. I can fax you some of our lab results if you like and some info on 9000 series concentrates.
IP: 209.244.85.97 |
Greg unregistered
|
posted 02-06-2003 05:24 PM
Update: At long last, I got to talk to the technician. My machine was worked on for the regular maintenance check and when it was returned to my station, this time, the conductivity was runnng 13.7- 14.0. The technician said that for all the previous maintenances he did not correct the high cond. alarm problem as he did not know there had been a problem with the high cond. alarm going off, and he had been told that the customer wanted the higher cond. of 14.2 - 14.5 as he became symptomatic at a lower cond.Well, this time, he took the liberty to lower the cond. range because he felt that the lower cond. was in limits for a 2K and thats what he had to do from his end despite my tx symptoms with lowered cond. What was causing the cond. to rise over 14.5 (bear with me as hopefully I will explain this correctly and if I don't you will still be able to see what the problem is) was a valve, or a valve behind the check valve was open all the way, instead of being at the quarter position it should of been at. This caused excessive pressure to come into the machine causing the cond. to rise too high setting off the alarms. I think he said this interfered with the machine pulling the concentrate rather than by positive pressure. So, he shut the valve back to the quarter postion ending the alarm situation. I don't recall what the technician said about why there were no alarms some days. Could it have something to do with pressure in the loop? He said the alarms should not have compromised my txs as by running the clock out I got the time back. But I told him several other technicians have said multiple alarms does compromise the tx and my clearances have been up and down ever since the problem started. He had no comment to that. I told him that I understood why he decided to handle the problem as he had, but now I had a new problem to solve, and after the tx was over I would be letting him know if I am symptomatic from the lowered cond. He gave me a way to get in touch with him in the future. I asked him if there was a way to increase the cond. rate of the machine with possibly another bicarb solution. He said if the bicarb solution was changed it would raise my potassium. So, he didn't seem to think there is any way this can be done. I had talked to the technican the first hour of tx. He left, and by the 2nd hour of tx I was already becomming symptomatic on the same machine I feel fine on when the cond. was 14-2-14.5. The cond. stayed around 14.0 most of the tx. I want to thank those of you who have stuck with me in trying to solve this problem. I had a feeling the cause for the high cond. alarms would have a simple explanation and it did. As you can see, the alarm situation is solved, but because I am now symptomatic with the lower cond. range, the problem is still not solved for me. I hope someone will have an answer for what I can do next as I will face one hurtful tx after another until this new problem is solved.
IP: 209.240.198.61 |
GTSCSS Member
|
posted 02-06-2003 05:47 PM
I think somebody doesn't REALLY know what the problems is. Unless he is talking about a 'pressure regulator' he shouldn't have to adjust any valves to get the right conductivity. If you are on a loop acid system (or bicarb or both) there should be check valves in the machine to keep the machine from being overwhelmed with too much pressure. Sounds to me like a pump spring is broken our about to break and letting too much acid in. I've had it happen before. Ask him if he checked that. As far as the symptoms, 14.0 is a good conductivity and I think you are just experiencing what all dialysis patients experience to some degree. I, however, feel for you and hope it gets better. Good luck!
IP: 65.215.145.67 |
Greg unregistered
|
posted 02-06-2003 07:56 PM
Something else I learned is bicarb for our unit is mixed using a combination of large bags and small boxes to get the amount needed. The mixing is done by machine. It says on the box of Naturalyte 4000: "Gently agitate the container a sufficient number of times to dissolve the powder. NOTE: Vigorous mixing and propeller style mixers can drive carbon dioxide from the solution and are therfore not recommended". So, since my unit is not heeding these instructions, what happens when carbon dioxide is driven from the solution or do you just find this a recommendation and not really a problem?
IP: 209.240.198.61 |
Greg unregistered
|
posted 02-07-2003 06:51 AM
GTSCSS, There were no alarms this tx. since the cond. is now at 14.0. The valve the technican spoke of is in the wall where the concentrate comes in. You're saying this valve being open all the way had nothing to do with the problem? He also said that in his regular maintenance he changed out the diaphrm, springs and seals.I am confused about your comment that all dialysis patients more or less feel unwell on dialysis. I didn't feel unwell the whole time my my machine's cond. was up at 14.3 -14.5. Thats many months of good txs. I don't know how cond. of 14.0 could be good for me if it makes me feel so sick. Nothing about the tx has changed but the cond. unless its an untrue reading. I will tell the technician what you said. Thanks for the good wishes.
IP: 209.240.198.61 |
RT unregistered
|
posted 02-07-2003 12:04 PM
If you are using central delivery system check the pressure in your bicarb and acid loop. If it's greater than 3psi you will have alot of conductivity problems. Good Luck, RT.
IP: 65.217.49.5 |
GTSCSS Member
|
posted 02-07-2003 12:09 PM
Like I said... if it's a 'valve' on the wall it shouldn't matter what position it's in. A combination of the pressure in the loop (provided it's within recommendations) and the check valves in the machine should keep the machine from having any conductivity problems. However, (like I said before) if it's a 'pressure regulator' he was adjusting then I could see how turning it closed a little more would help. Anyhow, as far as the conductivity goes, having a higher conductivity, theoretically, should cause an increase in clearances and fluid removal which generally causes patients to temporarily feel sicker while their body adjusts. However in your case, maybe it's the other way around. Your body might not be getting ENOUGH fluid removed and enough electrolyte exchange. You might want to look at running a 'Sodium Variation' program during your treatment so you don't feel sick. This will temporarily raise the conductivity of your dialysate during a portion of the treatment and may help your problem. Hope that helps and once again, good luck. quote: Originally posted by Greg: GTSCSS, There were no alarms this tx. since the cond. is now at 14.0. The valve the technican spoke of is in the wall where the concentrate comes in. You're saying this valve being open all the way had nothing to do with the problem? He also said that in his regular maintenance he changed out the diaphrm, springs and seals.I am confused about your comment that all dialysis patients more or less feel unwell on dialysis. I didn't feel unwell the whole time my my machine's cond. was up at 14.3 -14.5. Thats many months of good txs. I don't know how cond. of 14.0 could be good for me if it makes me feel so sick. Nothing about the tx has changed but the cond. unless its an untrue reading. I will tell the technician what you said. Thanks for the good wishes.
IP: 65.215.145.67 |
mgunsolus Member
|
posted 02-07-2003 02:25 PM
I agree ask about a variable sodium. Our Dr. uses a linear sodium which starts at 150 and goes down stablizing at 140 by the end of their treatment. I also agree about the loop pressure.We have ours regulated down to 4 psi on both the acid and bicarb. Our conductivity on our machines run 13.8 - 14.2. It may just take some time for your body to adjust to the rapid changes it has gone through the past few months. I would give it a month or so and let everything stablize in your body and then see how you feel. Also mention to your machine tech to check the acid and bicarb pump volumes and maybe recalibrate the conductivity cells in the machine. It is not a huge task to do this and won't take him but about an hour or so. We do this on every quarterly PM that we do on our machines. I am glad you got the problem worked out. You may also ask your Doctor about the way you feel now. He would be the best person to ask. Most of us just understand the technicial part of the dialysis process not the medical part. Therefore we can only guess as to what is going on with you personally. I am glad you finally got the anwsers you you seeked.
IP: 208.46.13.226 |
Greg unregistered
|
posted 02-07-2003 04:04 PM
mgunsoius, The technician said he checked the conductivity cells and they were good. Acid and bicarb pump volumes should be checked for what purpose?As patients involved with our txs, we do consult with our medical staff, but in jiffy lubes the training is minimal. My doctor and staff know very little about sodium modeling programs. The doctor's only solution is to take a patient .5 up or .5 down on his dry weight when the machine is capable of customizing the tx. Sodium modeling is an obvious possible solution, but its not obvious to them. I'm ready to proceed to this step, but will have to get info about it elsewhere as all my staff could do is set a program pot luck. I do not consider that safe medical care. I would like to be able to switch to another unit that has more competent care, but there are no others in my area. That is often the way it is for many patients. I'd also like to get into a home hemo program, but likewise, no programs in the area. Thanks for your continuous help and concern!
IP: 209.240.198.61 |
Chuck W Member
|
posted 02-07-2003 06:50 PM
It has become very apparent that Greg/Jane/patient's purpose is to spread the rhetoric of another website. I for one am finished posting replies to this individual.
IP: 152.163.207.194 |
Dennis Todaro Moderator
|
posted 02-07-2003 07:50 PM
I think we've reached one of those "Aha" moments in this discussion so I'll throw my 2 cents in.Here's your problem: pressure regulation in the concentrate supply loop is being controlled by partially closing the concentrate valve at the station. You are on a central feed for the acidified concentrate - machines are designed to draw concentrate via the concentrate pump and are not designed to accept concentrate under pressure. When the concentrate feed pressure exceeds 3 psi or thereabouts you can have a situation where the concentrate is forced passed the spring / check valve assembly on the concentrate pump causing momentary increases in the final conductivity. The older the pump spring / check, and the higher the pressure the more likely this will occur. This affect can be transitory dependent on how much concentrate is being taken from the feed piping by the other machines. (Pressure in the line decreases as more machines draw concentrate). Or it can be a matter of where your station is relative to the feed piping. (First on the line the higher the pressure). Since you state this generally happens toward the end of your treatment I would also guess that you dialyze on the last shift of the day or your treatments may run longer than others during that shift. As more machines (or their dialysate flow) are turned off the pressure in the concentrate line increases. The higher the pressure the greater the chance that the concentrate will force through the pump. So there is nothing wrong with the machine (in fact given the circumstances it's doing exactly what it is supposed to do by alarming) just "not the best design" for the concentrate feed loop.
IP: 12.213.12.110 |
Greg unregistered
|
posted 02-09-2003 02:43 AM
Dennis Todaro, The high conductivity alarms did not always go off towards the end of the tx. There was no pattern to it. The alarms went off anytime during the tx and some days they didn't go off at all.I don't know where I am on the loop, but my chair is located in the middle of a wing. I do run the last shift. I'm pretty sure my tx time is longer than most of the patients on my wing except for one. Two or three get off before me. Other than making the high conductivity alarms go off, how else should this situation have affected my txs? Did it, or did it not, compromise the txs? Update: With the first tx on my machine, since the high cond. alarm problem was resolved, not being a good one, I was very nervous about going to my next tx. When I got to tx., my regular machine was not there and a different machine was there in its place. The conductivity on this machine was similar to that of my, machine since the technican recalibrated it (13.7-14.0), so I prepared for the worst. To my great surprise, the tx at this lower cond. rate was problem free. What do I deduct from this? To recap, for many months now, my regular machine's cond. was at 14.3 - 14.5 and I had no problems. Anytime I was switched to another machine due to maintenance or a different scheduling, if the cond of the machine was lower than 14.2, I had a very symptomatic tx. So, what can be said about this? Is it possible that the acid and bicarb proprtioning mixture is not adequate in particular machines?
IP: 209.240.198.61 |
mgunsolus Member
|
posted 02-10-2003 10:29 AM
I agree with you chuck! Whoever Greg is it is starting to get stupid! I think that you greg just needs to talk with your doctor and clinic manager and quit going on and on with this post. I am also through with posting replies with this topic!
IP: 208.46.13.226 |
Enough unregistered
|
posted 02-10-2003 09:59 PM
Greg, Enough is enough!!!
IP: 67.26.49.160 |
Dennis Todaro Moderator
|
posted 02-11-2003 03:05 AM
Greg unfortunately the info you've provided while as concise as you could make it is not sufficient to make a definitive diagnosis of your problem. From what you've described the intermittant conductivity alarms probably had more to do with the concentrate loop pressure than any problem with machine function or settings. Since you stated somewhere along the line that you dialyze by the clock on the machine (rather than the clock on the wall) I would guess your treatment time is sufficient. The change in final conductivity should be based on the expected final conductivity of the formualtion used. If this is correct than the machine delivers the correct dialysate and alarms accordingly. As far as your feeling poorly this can be the result of many things which may or may not have anything to do with your treatment -certainly not the machine if it is functioning correctly. This is something which you need to discuss with your physician - I don't think you'll get the answer here. As far as comments ie - jiffy lube dialysis - well that's a perception. No one would deny that there are units out there where quality of care is a phrase and not a reality, however by and large most dialysis units deliver acceptable care - maybe your's is the exception - if so then you should address this with the facility's manager and medical director. Since this is the "technical" forum we should limit our discussion to technical topics - when we attempt to diagnose specific problems experienced by a patient we run the risk of misinforming. In our attempts to be helpful we may ultimately hinder a resolution. I won't close this topic yet but I don't think you'll get much more information than you already have.
IP: 12.213.12.110 |
GTSCSS Member
|
posted 02-12-2003 04:31 PM
quote: Originally posted by mgunsolus: I agree with you chuck! Whoever Greg is it is starting to get stupid! I think that you greg just needs to talk with your doctor and clinic manager and quit going on and on with this post. I am also through with posting replies with this topic!
I agree.... no more from me. This really should be dealt with at a clinical level. I wash my hands of it.